Laser's power supply unit of Biorad confocal microscope is broken - information is needed
#44
Author: adiraveh
Date: Tue, 16 Sep 2008 07:03
26 lines
2505 bytes
Hi all,

Our confocal microscope’s power supply unit (biorad mrc-1024 confocal)
is out of order. It is an Ion Laser Technology (ILT) unit, model 5403B
(from July 1995).

We would like to ask if any of you might have the electricity circuit
scheme for this unit (or a similar one). It would be very helpful for
us for trying to figure out the problem there.

Trying to find the problem in our electronics laboratory, we’ve found
that the unit can’t operate and get power without being connected to
the other units (as it is connected when it works on our whole setup).

However, we don’t have the interlock plug for the interface connector
that might probably solve this. We wonder if anybody could please
direct us whether there's a way to jumper the connector/circuit such
that we could operate the power supply at electrician place, where it
stands alone, without bringing the other units that to his place.


Many thanks in advance,

Adi,

The Weizmann Institute of Science.
Re: Laser's power supply unit of Biorad confocal microscope is broken - information is needed
#45
Author: Helpful person
Date: Tue, 16 Sep 2008 11:38
34 lines
2900 bytes
On Sep 16, 3:03 am, adira...@gmail.com wrote:
> Hi all,
>
> Our confocal microscope’s power supply unit (biorad mrc-1024 confocal)
> is out of order. It is an Ion Laser Technology (ILT) unit, model 5403B
> (from July 1995).
>
> We would like to ask if any of you might have the electricity circuit
> scheme for this unit (or a similar one). It would be very helpful for
> us for trying to figure out the problem there.
>
> Trying to find the problem in our electronics laboratory, we’ve found
> that the unit can’t operate and get power without being connected to
> the other units (as it is connected when it works on our whole setup).
>
> However, we don’t have the interlock plug for the interface connector
> that might probably solve this. We wonder if anybody could please
> direct us whether there's a way to jumper the connector/circuit such
> that we could operate the power supply at electrician place, where it
> stands alone, without bringing the other units that to his place.
>
> Many thanks in advance,
>
> Adi,
>
> The Weizmann Institute of Science.

Try asking Zeiss, they purchased Biorad's confocal microscope
division.


http://www.bio-rad.com/B2B/BioRad/offices/about/wn-press.jsp?BV_SessionID=@@@@1500455156.1221565003@@@@&BV_EngineID=ccchadefeihljlfcfngcfkmdhkkdfll.0&siteSection=about&selectedYear=2008&editorialOID=796232
Re: Laser's power supply unit of Biorad confocal microscope is broken - information is needed
#46
Author: Helpful person
Date: Tue, 16 Sep 2008 13:21
44 lines
3128 bytes
On Sep 16, 7:38 am, Helpful person <rrl...@yahoo.com> wrote:
> On Sep 16, 3:03 am, adira...@gmail.com wrote:
>
>
>
>
>
> > Hi all,
>
> > Our confocal microscope’s power supply unit (biorad mrc-1024 confocal)
> > is out of order. It is an Ion Laser Technology (ILT) unit, model 5403B
> > (from July 1995).
>
> > We would like to ask if any of you might have the electricity circuit
> > scheme for this unit (or a similar one). It would be very helpful for
> > us for trying to figure out the problem there.
>
> > Trying to find the problem in our electronics laboratory, we’ve found
> > that the unit can’t operate and get power without being connected to
> > the other units (as it is connected when it works on our whole setup).
>
> > However, we don’t have the interlock plug for the interface connector
> > that might probably solve this. We wonder if anybody could please
> > direct us whether there's a way to jumper the connector/circuit such
> > that we could operate the power supply at electrician place, where it
> > stands alone, without bringing the other units that to his place.
>
> > Many thanks in advance,
>
> > Adi,
>
> > The Weizmann Institute of Science.
>
> Try asking Zeiss, they purchased Biorad's confocal microscope
> division.
>
> http://www.bio-rad.com/B2B/BioRad/offices/about/wn-press.jsp?BV_Sessi...- Hide quoted text -
>
> - Show quoted text -

Sorry, the link to the news release does not work.  Cleverly designed
Biorad web site!  However, Zeiss should be able to help.
Re: Laser's power supply unit of Biorad confocal microscope is broken - information is needed
#47
Author: Kevin Cunningham
Date: Tue, 16 Sep 2008 13:44
44 lines
3233 bytes
On Sep 16, 3:03 am, adira...@gmail.com wrote:
> Hi all,
>
> Our confocal microscope’s power supply unit (biorad mrc-1024 confocal)
> is out of order. It is an Ion Laser Technology (ILT) unit, model 5403B
> (from July 1995).
>
> We would like to ask if any of you might have the electricity circuit
> scheme for this unit (or a similar one). It would be very helpful for
> us for trying to figure out the problem there.
>
> Trying to find the problem in our electronics laboratory, we’ve found
> that the unit can’t operate and get power without being connected to
> the other units (as it is connected when it works on our whole setup).
>
> However, we don’t have the interlock plug for the interface connector
> that might probably solve this. We wonder if anybody could please
> direct us whether there's a way to jumper the connector/circuit such
> that we could operate the power supply at electrician place, where it
> stands alone, without bringing the other units that to his place.
>
> Many thanks in advance,
>
> Adi,
>
> The Weizmann Institute of Science.

Adi,

You can make that jumper from parts in a good electronics catalog and
it shouldn't cost you much.  Try these (www.mouser.com, www.alliedelec.com,
www.jameco.com). You might have to buy a plug and some cable and then
close the right lines but that is much cheaper than Zeiss's price.

When you find the messed up component(s) the catalogs should help you
get a replacement.  There is a problem though, some times components
are custom designed and programmed.  This can be a real headache, I've
had to tell people that I found the problem and they don't make the
solution any more.  Then life gets real interesting.....

Thanks,

Kevin Cunningham
SMS
Re: Laser's power supply unit of Biorad confocal microscope is broken - information is needed
#48
Author: adi
Date: Tue, 16 Sep 2008 14:33
71 lines
4106 bytes
On Sep 16, 4:44 pm, Kevin Cunningham <sms...@mindspring.com> wrote:
> On Sep 16, 3:03 am, adira...@gmail.com wrote:
>
>
>
>
>
> > Hi all,
>
> > Our confocal microscope’s power supply unit (biorad mrc-1024 confocal)
> > is out of order. It is an Ion Laser Technology (ILT) unit, model 5403B
> > (from July 1995).
>
> > We would like to ask if any of you might have the electricity circuit
> > scheme for this unit (or a similar one). It would be very helpful for
> > us for trying to figure out the problem there.
>
> > Trying to find the problem in our electronics laboratory, we’ve found
> > that the unit can’t operate and get power without being connected to
> > the other units (as it is connected when it works on our whole setup).
>
> > However, we don’t have the interlock plug for the interface connector
> > that might probably solve this. We wonder if anybody could please
> > direct us whether there's a way to jumper the connector/circuit such
> > that we could operate the power supply at electrician place, where it
> > stands alone, without bringing the other units that to his place.
>
> > Many thanks in advance,
>
> > Adi,
>
> > The Weizmann Institute of Science.
>
> Adi,
>
> You can make that jumper from parts in a good electronics catalog and
> it shouldn't cost you much.  Try these (www.mouser.com,www.alliedelec.com,www.jameco.com). You might have to buy a plug and some cable and then
> close the right lines but that is much cheaper than Zeiss's price.
>
> When you find the messed up component(s) the catalogs should help you
> get a replacement.  There is a problem though, some times components
> are custom designed and programmed.  This can be a real headache, I've
> had to tell people that I found the problem and they don't make the
> solution any more.  Then life gets real interesting.....
>
> Thanks,
>
> Kevin Cunningham
> SMS- Hide quoted text -
>
> - Show quoted text -

Dear Kevin,

Many thanks for your good ideas.

The thing is that for regular use, the interface connector is simply
connected to the computer and power supply gets voltage without any
need of interlock plug or jumper.

Now, as we had to fix the power supply unit, we took it alone to the
electrician, and we need a way to activate it without the other units,
and we don't have an interlock plug. The worse thing is that we don’t
even know what the right lines to close are, since we don’t have any
scheme of the electric circuit...

But thanks anyway

Adi
Re: Laser's power supply unit of Biorad confocal microscope is broken - information is needed
#49
Author: adi
Date: Tue, 16 Sep 2008 14:55
70 lines
3732 bytes
Dear Helpful person,

Many thanks,

I talked with Zeiss. However, as I was told, this ILT power supply
unit is old and its brand is not longer exist (for more than 10
years). He will make efforts to try and get a scheme, but he can not
promise he will succeed…
I thought that maybe somebody here might have one.

Thanks again,

Adi

On Sep 16, 4:21 pm, Helpful person <rrl...@yahoo.com> wrote:
> On Sep 16, 7:38 am, Helpful person <rrl...@yahoo.com> wrote:
>
>
>
>
>
> > On Sep 16, 3:03 am, adira...@gmail.com wrote:
>
> > > Hi all,
>
> > > Our confocal microscope’s power supply unit (biorad mrc-1024 confocal)
> > > is out of order. It is an Ion Laser Technology (ILT) unit, model 5403B
> > > (from July 1995).
>
> > > We would like to ask if any of you might have the electricity circuit
> > > scheme for this unit (or a similar one). It would be very helpful for
> > > us for trying to figure out the problem there.
>
> > > Trying to find the problem in our electronics laboratory, we’ve found
> > > that the unit can’t operate and get power without being connected to
> > > the other units (as it is connected when it works on our whole setup).
>
> > > However, we don’t have the interlock plug for the interface connector
> > > that might probably solve this. We wonder if anybody could please
> > > direct us whether there's a way to jumper the connector/circuit such
> > > that we could operate the power supply at electrician place, where it
> > > stands alone, without bringing the other units that to his place.
>
> > > Many thanks in advance,
>
> > > Adi,
>
> > > The Weizmann Institute of Science.
>
> > Try asking Zeiss, they purchased Biorad's confocal microscope
> > division.
>
> >http://www.bio-rad.com/B2B/BioRad/offices/about/wn-press.jsp?BV_Sessi...Hide quoted text -
>
> > - Show quoted text -
>
> Sorry, the link to the news release does not work.  Cleverly designed
> Biorad web site!  However, Zeiss should be able to help.- Hide quoted text -
>
> - Show quoted text -

Dear
Re: Laser's power supply unit of Biorad confocal microscope is broken - information is needed
#50
Author: osr
Date: Tue, 16 Sep 2008 16:04
5 lines
1703 bytes
DZ Laser, www.dzlaser.com

now supports ILT

Steve